An image demonstrating motion and adequate density was obtained using 100 mA at 0.5 seconds. If the time is changed to 0.25 seconds, what mA should be used to maintain density?
A) 100 mA
B) 200 mA
C) 300 mA
D) 400 mA
Correct Answer:
Verified