A pulmonary function technologist uses an isothermal lung analog to check the accuracy of a constant volume plethysmograph.The lung analog has a volume of 4.00 L.Simulating the VTG maneuver produces the following volumes: On the basis of these findings,the technologist should:
A) Check the door seal for leaks or cracks.
B) Recalibrate the box pressure transducer.
C) Zero the mouth pressure transducer and repeat the measurement.
D) Begin patient testing with the body box.
Correct Answer:
Verified
Q30: Quality control is performed on a cartridge-based
Q31: According to the current ATS-ERS guidelines,spirometers should
Q32: A 3-L syringe is used to
Q33: Spirometers used to measure FEV1 should be
Q34: A patient whose FVC is 4 L
Q36: According to 2005 ATS-ERS recommendations,calibration syringes should
Q37: A DLCO simulator using two precision gases
Q38: The calibration gas for a blood gas
Q39: The acceptable range for accuracy of a
Q40: A patient whose VC is 3.5 L