Spirometers used to measure FEV1 should be able to measure flows in the range of 0 to 14 L/sec (ATS-ERS,2005) .The maximal resistance allowable at all flows is:
A) Less than 6 cmH2O/L/sec
B) Less than 1.5 cmH2O/L/sec
C) Determined by the type of filter used
D) One half of the flow rate in L/sec
Correct Answer:
Verified
Q28: A biologic control subject performs DLCO measurements
Q29: Quality control runs on a PCO2 electrode
Q30: Quality control is performed on a cartridge-based
Q31: According to the current ATS-ERS guidelines,spirometers should
Q32: A 3-L syringe is used to
Q34: A patient whose FVC is 4 L
Q35: A pulmonary function technologist uses an isothermal
Q36: According to 2005 ATS-ERS recommendations,calibration syringes should
Q37: A DLCO simulator using two precision gases
Q38: The calibration gas for a blood gas