Quality control runs on a PCO2 electrode have produced a mean of 40 mm Hg ± 2 mm Hg (1 SD) .What are the chances of obtaining a value of 45 mm Hg on a control run if the electrode is functioning properly?
A) 95%
B) 50%
C) 5%
D) Less than 1%
Correct Answer:
Verified
Q24: A patient performs spirometry,and the following results
Q25: A 3-L syringe is used to calibrate
Q26: A subject performs spirometry,and these results are
Q27: The greatest risk of cross-contamination in the
Q28: A biologic control subject performs DLCO measurements
Q30: Quality control is performed on a cartridge-based
Q31: According to the current ATS-ERS guidelines,spirometers should
Q32: A 3-L syringe is used to
Q33: Spirometers used to measure FEV1 should be
Q34: A patient whose FVC is 4 L